How to convert pixels in IFCB images to micrometers?

To determine sizes of plankton and particles imaged by the Imaging FlowCytobot (IFCB), one needs to convert pixels to units of length.

At the OOI Facilities Board IFCB Focus Group meeting in June 2025, data exercises used the following conversion factor for OOI’s IFCB images:
2.77 pixels per micrometer

This factor is appropriate in practice for recent IFCB serial numbers (note older IFCBs may have different conversion factor).

The McLane IFCB Datasheet (version 12/22) says “∼ 2.7 pixels/micron” because the precise factor should be determined for each instrument.

To confirm the factor 2.77 for OOI’s serial numbers 199, 200, and 206, we recommend evaluating several beads runs. The beads are Invitrogen “AlignFlow Plus 6 um” with actual size 5.6 um.

As a quick check: if 2.77 px/µm is correct, a 5.6 µm AlignFlow bead should measure about 15.5 px across. If your segmented bead diameters cluster near that, it supports the factor for serials 199, 200, and 206. Use equivalent circular diameter and a consistent segmentation threshold, since edge thresholding can bias small bead measurements.